Is it possible to select the external equipment on Teseq NSG 4060

Discussion and questions regarding the RadiMation device drivers that control the measurement equipment.
Post Reply
kunan
Posts: 4
Joined: 10 Jan 2022, 15:37

Is it possible to select the external equipment on Teseq NSG 4060

Post by kunan »

Hello,

to switch automatically between measuring equipment for EN 61000-4-16 and EN 61000-4-19 one can use the external switch SW4060. The position of the switch is selected using the Equipment commands as described in the manual. (EQUIPment:SELect CDND, EQUIPment:SELect CT, EQUIPment:SELect NSG4060-1)

Is it possible to configure this commands in Radimation?
Jeru
Posts: 44
Joined: 09 Jun 2020, 15:48

Re: Is it possible to select the external equipment on Teseq NSG 4060

Post by Jeru »

Hello Kunan,

This is possible by including a configurable switchmatrix driver in the test-site that you are using for each measurement type.

The configurable switchmatrix driver can be configured to send user defined commands to connected devices when certain events are triggered in the RadiMation software. By doing so it is possible to for example send the equipment select commands to the NSG4060 moments before the test is started.

For more information on configurable switchmatrix driver please find our wiki @ https://wiki.radimation.com/wiki/index. ... tch_Matrix.


I hope to have informed you,
kunan
Posts: 4
Joined: 10 Jan 2022, 15:37

Re: Is it possible to select the external equipment on Teseq NSG 4060

Post by kunan »

Hello Jeru,
works perfect!

Thank you