How to solve SRQ line errors

Posted: 27 May 2014, 08:46
by solutionseeker
During start-up we got an error that a SRQ line was raised. How can be avoid this error being generated?

Re: How to solve SRQ line errors

Posted: 27 May 2014, 08:52
by sast
To solve any issues with a detected SRQ line error, it is first needed to locate the device which is raising the SRQ on the bus.

1. Start RadiMation
2. Yes, SRQ error is shown? Shutdown 1 device, => 1.
3. No, SRQ error is not shown? Most likely the device which was shutdown in the previous step was the problem.

Sometimes, disconnect and blow in the GPIB connections might also solve these kind of errors.

When this error occurs on regular base, please contact the RadiMation support.

See also the description in the Knowledgebase: http://wiki.dare.nl/wiki/index.php/Radi ... _805306952 and http://wiki.dare.nl/wiki/index.php/Radi ... _805306953.