Is the correct calculation for the Uniform Field Area used?

Posted: 30 Aug 2011, 21:30
by joro
Is RadiMation® using the correct method for the Uniform Field Area calibration?

Re: Is the correct calculation for the Uniform Field Area us

Posted: 30 Aug 2011, 21:30
by joro
The latest EN 61000-4-3 standard describes two methods to perform a "field homogeneity verification" (also called 'Uniform Field Area calculation') (the constant field method and constant power method). RadiMation® supports the constant field method while this will result in the highest accuracy (With this procedure, the linearity error of the field probe is not influencing the calibration accuracy. While the linearity of a power meter is much better than the linearity of a field probe, this method is, from an accuracy point of view, the better solution). Although RadiMation® does not support both methods, this does not mean RadiMation® can not be used for field homogeneity verification, because according to the standard, both methods should result in the same calibration file.