We have released a new version of the RadiMation software, which includes several improvements and fixes for problems that have been detected since the release of the 2016.2.10 version.
This version 2017.1.4 has been validated by our internal software quality checks and also has passed validation at DARE!! Measurements. We strongly advise to test and verify (according to your quality system) this and each new version of RadiMation at for example a separate PC or laptop with your setup and with your measurement equipment, before you are going to use it for daily operation.

Download
The new 2017.1.4 release of RadiMation can now be downloaded from the DARE!! web-site. Image

We always advise to thoroughly test EACH new RadiMation software version to determine if it works correctly in your situation with your equipment, before you use it for daily measurements. This is even a requirement for all ISO 17025 accredited test-labs. To enable customers to validate new software versions before they implement that new software version, we even made it possible to install multiple versions of RadiMation side-by-side on a single PC. It is of course also possible to perform this validation on for example a separate PC or laptop.

Known issues
We already are aware of 1 known issue in this version.
  • It is possible that the password of a newly created engineer cannot be changed.
If you have a problem or question, please do not hesitate to report it to us, so we can continue to improve RadiMation.

New functionality
We also include several improvements and new functionalities in version 2017.1.4, the most important improvements are:
  • Graphs filtering in multiband immunity, which fixes the overwriting graph problem
  • 1 Hz on/off software pulse modulation
  • New user interface for Sequence configuration window.
  • Several user interface and usability improvements.
  • Report generator fixes.
  • Better control of measurement equipment.
Fixed issues
The following issues are fixed in this 2017.1.4 version (compared to the 2016.2.10 version):
  • #8399: Loading a multiband immunity test is not possible, unable to create VSWRGatherer
  • #8397: Searching for a customer in the customer database closes the window before the customer can be selected
  • #8389: Received power is rounded in Antenna Scan EUT Calibration
  • #8383: Store RADIMAT_SESSION.LOG in a PC local temporary folder
  • #8382: ODBC customer database is not working anymore
  • #8301: Unhandled exception access to the path ErrorModeFile.radilog is denied
  • #8244: Climat typo it should be Climate
  • #8210: Opening a RE Multiband test is taking one and a half minutes
  • #8363: RadiLog option in view menu incorrect behaviour
  • #8238: Language file 0.lng
  • #8199: Opening Multiband Immunity results takes a lot of time since graph filtering is implemented
  • #8145: Unable to show AD Channels in 2017.1.1
  • #8097: Linear and logaritmische scale setting are not saved
  • #8091: Improve calculation of multiband immunity graphs
  • #8045: RDC files are always stored to the Global configuration path while standalone setting is used
  • #8144: Exporting graph to LLF in the graph options crashes
  • #8140: VI Signal Generator all controls are still disabled after connected
  • #8138: VDI Impedance is set incorrectly to the device
  • #8137: Language files as unicode
  • #8130: Virtual Instruments infinit loop errors
  • #8114: Calibration Files do not use the same calibration method
  • #8112: Radilog Parsing error when opening log from zip
  • #8109: Export full PEAK table
  • #8108: Error window tooltip causes object reference error.
  • #8107: Errors are not reported in RadiMation
  • #8106: RadiLog object reference after saving
  • #8104: RadiMation does not automatically refer and open the latest installed version
  • #8101: RadiLog SQL files containing ";" is not opening correctly
  • #8096: Auto detect peaks shows object reference error
  • #8085: Gaps in graph lines.
  • #8069: Shortcutkeys not updating ctlEvents
  • #8061: Before actions are prompted multiple times at the end of the test
  • #8054: Manual frequency is not measured when it is exactly the same as the last frequency in the calibration file.
  • #8053: EUT being removed after network reset.
  • #8046: Error report, cant write to temporary directory
  • #8040: Error window location cut off
  • #8033: Combine user defined units to prevent multiple equal axis from being shown in graph
  • #8030: VI powermeter windows does not work correctly
  • #8029: Graphs during Multiband Immunity are not updated during test
  • #8021: When a Band has been selected the TSF is marked as being changed
  • #8015: Improve graphical representations of graph
  • #8011: Exporting polar graphs contains a white space area.
  • #8009: Opening a second CE Absorbing clamp test, switches both graphs from log to linear xaxis.
  • #8004: Opening of locked files should wait for a few seconds and retry opening locked files
  • #8003: Clear graph button for VI windows does not work (when disconnected)
  • #7980: Test data files are zero based count on disk which is confusing for end-users
  • #7971: Radilog saving file, file in use while reporting error.
  • #7969: EFT tests with repetition rate 100 kHz (burst duration 0,75ms) can NOT be performed
  • #7966: Last View update is being skipped
  • #7965: A read only TSF is also asked to save the changes
  • #7964: Directly after installation, the power graphlines are not shown in the VI Powermeter
  • #7963: RadiLog parsing fail
  • #7962: RadiLog to long subject gives error. Strange behaviour occurs.
  • #7955: Size of Voltage Dips and interrupts TSF window can be made too small
  • #7953: Frequency list settings are not saved for a sequence and thus not used when Fixed power test is run from within a SEQ file
  • #7951: Sending unprintable characters with configurable device driver
  • #7950: Ability to hide Report Generator Template column for 'The keyword "" is available in the tempate' window
  • #7947: Switching to a different band using 'Continue automatic mode settings' is incorrectly checking if the carrier frequency is valid
  • #7946: Opening test results directly sets the changed flag.
  • #7940: Why is the Peak Table disabled during an Emission test?
  • #7937: Signal generator VI frequency can be set outside the range of the controlled device
  • #7932: Very long filenames of the RadiLog file cannot be opened by RadiLog
  • #7927: Duplicate more than 1 test at the same time in the new Sequence window.
  • #7926: The RBW value has changed by the device from 120 to 100 User description:
  • #7922: External modulation should not be possible for a VI signal generator
  • #7918: Graph customization window can be made too small
  • #7915: Merging of calibration files looses the originally desired fieldstrength
  • #7908: Invalid tsf in sequence can corrupt the sequence
  • #7905: ReportGenerator Report and output file selection is not saved.
  • #7904: Default table settings are stored in the user interface
  • #7903: Unable to start RadiMation with incorrectly configured path
  • #7902: MessageBox that is shown from the testsite before actions can be shown behind the main window
  • #7898: & sign in device driver name is not correctly shown in the EUT Monitoring dialog
  • #7897: Error Report send not always sending.
  • #7896: Missing last test frequency in ReportGenerator output
  • #7894: Radilog memory is not being swapped
  • #7893: Data file is not saved every 5 minutes, when a final peakmeasurement is busy
  • #7887: Missing keyword in Table shifts the table
  • #7886: Resizing of the Graph customization window hides some graph customization controls
  • #7885: Test data like notes and specific templates settings are updating or overwriting EUT overview settings
  • #7878: RadiLog error report containing # in description gives error.
  • #7867: Out of memory when starting RI test
  • #7866: Corrupt file saving when validator failed.
  • #7845: PI test EUT cable selection should be changable while TSF is read only.
  • #7844: Incorrectly saved Test results starts the test when reviewing the Test results
  • #7842: Saving EUT file with keywords
  • #7836: Connect button is not showing 'Connecting'
  • #7824: Reflected power input in Multiband Immunity
  • #7820: Test specific templates are gone after graph customization
  • #7819: No data document could be created for the test identifier: "
  • #7812: VI Fieldsensor keeps raising error messages when the probe is turned off before it is disconnected
  • #7811: Clear button on the VI Fieldsensor doesn't seem to work when the table is shown
  • #7799: Sorting of the ctlFormattableGrid is different when new columns is added
  • #7796: Compression Gain System Calibration default 1 amount of samples seems to result in a corrupt calibration file
  • #7779: Overlapping bands in MB immunity are tested multiple times when only the modulation is different
  • #7773: Incorrect table information for frmVIPowerMeter at the ctlControlGraph table
  • #7769: Saving tsf as ReadOnly data lost
  • #7768: Wrong frequency at monitoring input events
  • #7760: Ability to merge multiple Radiated Immunity CALibration files for different frequency ranges into one
  • #7737: Report generator query keyword for the table information is not exporting any data
  • #7721: System attenuation gain duplicate measurements
  • #7717: Unchanged LLF is asking to be saved when closing the window.
  • #7644: Filling the EUT information is being slow.
  • #7637: Driver: emtest UCS500-M4
  • #7581: Test description of the test is not visible in the test data
  • #7575: RadiLog long tests multiple log files.
  • #7523: PI EFT Test levels are not covering all the different measurements
  • #7499: RadiMation opening EUT is slow
  • #7449: VI Powermeter error lock when removing device
  • #7441: Minimum Form Size is always set to 0,0 at runtime
  • #7411: Connection state is still orange while the device is already connected
  • #7365: Minimum size of 'Burst' and 'Envelope Tracing' tabs of the VI Powermeter
  • #7255: Change Basic small controlling window for Amplifier
  • #7254: Change Basic small controlling window for PowerMeter
  • #7252: RadiMation Free Responsiveness
  • #7249: Show example of unit settings
  • #7247: Prana amplifier is not correctly being put in Operate during immunity test
  • #7242: Two electrical field graphs are shown. Which one is correct?
  • #7214: PI windows still do not fit at full window size using an acceptable resolution
  • #7203: A generic error occurred in GDI+
  • #7177: Unable to zoom graph while test is running
  • #7167: Amplifier linearity measurement accepted power lines hidden
  • #7147: Redoing an already performed test overwrites the previous test-results
  • #7077: continuous measure is using sweep time setting
  • #7074: Possibility to change polarization of an existing calibration file
  • #6907: Improve autosave functionality that it no longer overwrites the current test data, but first makes a backup
  • #6861: Radiated Immunity Multiband is referring to a receiver
  • #6452: Frequency array calculation of emission sweep of analyzer is missing a step
  • #6449: Frequency range settings are generating an error and the configuration of the TSF is lost when a read only test is saved as new test
  • #6352: Changing the activated detector during Continuous Measure clears the measured value of the deactivated detector
  • #6264: Overlapping bands in MB immunity are tested multiple times when only the modulation is different
  • #6230: Table export only copies the time and frequency data
  • #6191: Add an 'Add Test' button to the Sequence window, to simplify the addition of a test
  • #6075: During instable measurements it may happen that Peak and Quasi-Peak measurements are inconsistent
  • #5545: Graph is not updated when switching to the second band
  • #5529: Invalid Device Driver Handle (Virtual) Amplifier
  • #5469: Invalid Device Driver Handle error (Kalmus 727LC)
  • #5440: Can't zoom out after maximum zoom in at graph
  • #5341: After a Manual Mode change there is undesirable regulation
  • #5307: Despite of no changes the message "The file has changed, do you want to save changes?" is shown
  • #5216: Test started event log entry should only be added in a restarted multiband test if manual or automatic is pushed
  • #5189: Sweeptime is used during final measurement instead of the QP measurement time
  • #4248: Error Multiband start and end frequency
  • #4159: RBW in peak table (RE multiband)
  • #4001: Difficulties to differ results from 2 bands with the same frequency range in multiband immunity tests
  • #3970: Radiated emission multiband: detector from pre-measurement used during "continuous measure"
  • #3598: Configurable Peak Table
  • #3241: RI multi band test is overwriting the previous information
  • #3032: Opening an Multiband test while performing a Multi band test is messing up the test
  • #2947: Amplifier is turned off when closing another multiband test
  • #506: 1 Hz on off modulation is not possible according to the DO-160E / ED-14E standards
If you have specific questions regarding on of the above items, please contact us at support@radimation.com