(#7577) EUT calibration one *cor file

Posted: 22 Sep 2016, 12:20
by heutink
RF generator RadiGen RGN6000B has two RF outputs. To do an EUT calibration over the whole range of both generator outputs, two tests have to be configured and put together via a sequencer to perform in one run. Each test can generate its own (frequency sub-range) *.COR file.

Is there an other (fast) method to combine both independant *.COR files to one new *.COR file without "copy to" and "paste from" clipboard actions.

Re: EUT calibration one *cor file

Posted: 26 Sep 2016, 12:17
by sast
We assume that the EUT calibration is performed to determine an output correction of both RF outputs.
The current method you suggested is the most effective way to have it done, because both frequency ranges are generating their own correction of that range.
You can indeed combine these results in the way you have described.