How to set Test level Limitation Multiply factor Multiband

Posted: 15 Nov 2013, 16:50
by solutionseeker
When we want to perform a Conducted Immunity Fixed Current test it is possible to make use of a Multiply factor for the Test level Limitation, as can be seen from the screenshot below.
CI_SB_Fixed_Current_HowToSet_MultiplyFactor.png
CI_SB_Fixed_Current_HowToSet_MultiplyFactor.png (48.21 KiB) Viewed 7054 times
How can we set the configuration of this setting using the Multiband test?

Re: How to set Test level Limitation Multiply factor Multiba

Posted: 15 Nov 2013, 16:55
by sast
This is possible by opening the Conducted Immunity Multiband test and Click the "Add" button under the "Test Level" list. A new window with a list of Test Levels and Limits will be shown.
Now select from this list the Limit "Calculated Current Limit". The following window will be opened:
CI_MB_Fixed_Current_HowToSet_MultiplyFactor.png
CI_MB_Fixed_Current_HowToSet_MultiplyFactor.png (24.19 KiB) Viewed 7052 times
In the Test Level box it is possible to fill in a "Constant Current' which can be x times the desired Test Level. It is also possible to fill in a x amount of "Calibrated Power".
This will correctly limit the Test Level used during the test.