Calibration on voltage level

Posted: 19 Jun 2020, 12:21
by heutink
Performing a "Calibration > System Calibration > Conducted Immunity" on a voltage test level. The only measuring instrument i use is a DVM. However, the software asks for sensor power meter a.s.o. How do i have to define this calibration on voltage level?

Re: Calibration on voltage level

Posted: 22 Jun 2020, 12:53
by Jeru
According to the standard that is used for testing different system calibration modules can be used.

If testing is done for most CE and Automotive purposes:
Calibration > System Calibration > Conducted Immunity can be used.

For this type of calibration it is not possible to use an multi-meter as device to regulate the voltage test levels.
A power meter represented by the sensor power meter is required for this type of calibration.

If testing is done for Mil standard purposes:
Calibration > System Calibration > Conducted susceptibility can be used.

For this type of calibration it is possible to use an multi-meter to regulate the voltage test levels.