What is the easiest way to correct multiple cables between the powermeter and the calibration jig?
Posted: 17 Jul 2020, 08:32
In our calibration setup for the Conducted immunity IEC 61000-4-6, we have multiple cables between the calibration jig and the analyzer that we are using to measure the injected voltage level.
At present I am having to combine all 5 transduced factors into one for the “Power Correction” when using the addressing the Sensor Power Meter (Spectrum Analyser).
In order to achieve the desired test severity level, do all the factors related to the monitoring (i.e. 40dB attenuator, cable 21, cable21a, cable701 and 30dB attenuator) need to be combined into one factor file and inserted in the “Power Correction” tab at the bottom of the Sensor power meter or is there another way of doing this?
As you can see from the image below we are using a Spectrum Analyser instead of a Sensor Power Meter, however there are more than one correction file we need to consider. They include the 40dB attenuator and Cable#06 outside the shielded enclosure and cables #21 a along with a 30dB attenuator inside the shielded enclosure.
It would be ideal if RadiMation could simply add additional correction files within the Correction file tab, however it appears that I can only add one factor file, hence why I have combined all the factor files together into a single file name.
How would you best recommend I address this issue?
At present I am having to combine all 5 transduced factors into one for the “Power Correction” when using the addressing the Sensor Power Meter (Spectrum Analyser).
In order to achieve the desired test severity level, do all the factors related to the monitoring (i.e. 40dB attenuator, cable 21, cable21a, cable701 and 30dB attenuator) need to be combined into one factor file and inserted in the “Power Correction” tab at the bottom of the Sensor power meter or is there another way of doing this?
As you can see from the image below we are using a Spectrum Analyser instead of a Sensor Power Meter, however there are more than one correction file we need to consider. They include the 40dB attenuator and Cable#06 outside the shielded enclosure and cables #21 a along with a 30dB attenuator inside the shielded enclosure.
It would be ideal if RadiMation could simply add additional correction files within the Correction file tab, however it appears that I can only add one factor file, hence why I have combined all the factor files together into a single file name.
How would you best recommend I address this issue?