ESPI 3 R&S receiver Reference level

Posted: 08 Oct 2012, 10:29
by Rosu
Hello John,

we use the ESPI3 receiver from R&s with option FSP-B9. The reference level is always set to 80dbµV. Can I change this value? I want to see the measured spectrum also on the receiver display.

Kind Regards,
Marius

Re: ESPI 3 R&S receiver Reference level

Posted: 10 Oct 2012, 09:49
by joro
Hello Marius,

Are you using the R&S ESPI3 in 'Spectrum Analyzer' or in 'Scanning Receiver' mode?

With best regards,
John

Re: ESPI 3 R&S receiver Reference level

Posted: 10 Oct 2012, 10:11
by Rosu
Hi John,

The most applications are in scanning mode, but some aplications also in swepped mode.

Greetz
Marius

Re: ESPI 3 R&S receiver Reference level

Posted: 10 Oct 2012, 10:55
by joro
When the R&S ESPI3 is used in Spectrum analyzer (Swepped) mode, the to be used Reference level can be configured by the 'Reference Level' setting in the TSF file of the Emission test.
When the R&S ESPI3 is used in Scanning Receiver (scanning) mode, the top of the graph is not controlled by RadiMation. The top of the graph is determined and set by the firmware of the R&S ESPI.

Does this clarify the observations you are seeing?

With best regards,
John

Re: ESPI 3 R&S receiver Reference level

Posted: 10 Oct 2012, 11:09
by Rosu
Hello John,

it seems to be so. I will check this asap.

Thx
Marius