Receiver settings stepsize and measure time are greyed out

Discussion and questions regarding the Radiated Emission and Conducted Emission modules of RadiMation.
Post Reply
Rdammert
Posts: 6
Joined: 09 Jun 2020, 15:46

Receiver settings stepsize and measure time are greyed out

Post by Rdammert »

I want to perform an emission test but i am not able to configure the measurement time and stepsize.
How can i configure the measurement time and stepsize of my test?
Jeru
Posts: 38
Joined: 09 Jun 2020, 15:48

Re: Receiver settings stepsize and measure time are greyed out

Post by Jeru »

When the measurement time and step size controls are greyed out, it means that the analyzer defined in the test site is not used in Scanning receiver or Time domain mode.

To have the measurement time and step size controls active, make sure you add either the Scanning receiver or Time domain version of the driver used to control your analyzer.

These specific version of the device driver can be added to the list of available devices just as any other new device driver to be added for use in RadiMation.

After adding the correct driver to the test site at the spectrum analyzer device slot, the controls should be made active again.
User avatar
joro
Posts: 440
Joined: 24 Aug 2011, 09:55
Contact:

Re: Receiver settings stepsize and measure time are greyed out

Post by joro »

Also see https://forum.radimation.com/viewtopic.php?f=6&t=881 , which is a similar question referring to a greyed out VBW and sweep time.
Post Reply