Page 1 of 1

Doing a Conducted emission measurement with a CurrentProbe

Posted: 30 Jul 2019, 14:48
by tjsc
When I look in the menu for conducted emission I only see that there is a single band test to do conducted emission tests with a current probe. Is it also possible to do a conducted emission test with a current probe in multiband?

Re: Doing a Conducted emission measurement with a CurrentProbe

Posted: 30 Jul 2019, 17:21
by tjsc
Edit: see below for an updated explanation.


Yes. That is correct. At this time there is no separate test for multiband. With the tools that are available it is possible to set this up and do multiband tests with the probe if needed.

We can use the Conducted Emission test as it essentially does the same, only the CE test assumes a LISN. If we can make a driver for a LISN so that the test thinks there is a LISN in the test but is actually reading the probe we have achieved our goal and have a multiband test setup for CE with a current probe.

Let's first make a driver for the probe to function as a LISN in the test:

Go to "Configuration/configuration" and open "Device Drivers"
Go to "LISN"
Add "Configurable LISN"
CofigurableLISN.PNG
CofigurableLISN.PNG (34.88 KiB) Viewed 363947 times
Rename the LISN driver to a appropriate name like: CurrentProbe/LISN
rename.PNG
rename.PNG (33.25 KiB) Viewed 363947 times
Set the frequency range.
The rest of the settings can be left blank.

Close and save the configuration.

Before we move on lets make sure tat we add our correction file that is normally used with the current probe driver.

Go to your current probe and check which file has been used for correction
ProbeCorrectionFile.png
ProbeCorrectionFile.png (68.13 KiB) Viewed 363947 times
Now go back to our LISN driver we just made and add the correction file there. This is very important as the correction is needed for the probe to measure correctly.
CuurentprobeLISN Correction.png
CuurentprobeLISN Correction.png (59.53 KiB) Viewed 363947 times
Now our driver is set up correctly. Let's now set up the testsite whith which we can do the tests.

Add a new testsite:
NewTestSite.png
NewTestSite.png (42.05 KiB) Viewed 363947 times
Add your amplifier, analyser/receiver and Current Probe/LISN driver. As the Current probe is now using a LISN driver you need to add it as a LISN.
Your test site should look something like this:
DevicesTab1.PNG
DevicesTab1.PNG (20.32 KiB) Viewed 363947 times
DevicesTab2.PNG
DevicesTab2.PNG (19.63 KiB) Viewed 363947 times
Close and save the test site.

Now we have setup the driver and made a new test site where it is added with all the other devices needed for this test.
The setup is now done and we can make a test.

Make a new EUT and go to Tests. Select Conducted Emission -> Multiband
This will open the test set-up window. As you can see, the settings are the same as a normal CE test.
It is important that you turn on at least one of the lines, the easiest is to just check Nutral.
TestCurrentProbe.png
TestCurrentProbe.png (53.26 KiB) Viewed 363947 times

Re: Doing a Conducted emission measurement with a CurrentProbe

Posted: 13 Nov 2025, 15:42
by PhilipM
I came here looking for this information,

I would prefer not to set up a Current Clamp as a LISN. I am looking at performing CISMR 25 testing.

Re: Doing a Conducted emission measurement with a CurrentProbe

Posted: 13 Nov 2025, 15:51
by PhilipM
These are my screen shots

Re: Doing a Conducted emission measurement with a CurrentProbe

Posted: 13 Nov 2025, 17:10
by roge
Dear Philip,

To perform a multiband conducted emission measurement with a current probe, create a new TSF file by choosing "Tests" > "Conducted emission" > "Multiband". When configuring the test make sure to select current sensor as location type, this ensures that the values in the test result are shown in 'current'.
CEMBCurrentProbe.png
CEMBCurrentProbe.png (51.78 KiB) Viewed 337959 times
Additionally the current unit can be changed by clicking the 'Units' button: