Possible to use 4 probes during field calibration?

Discussion and questions regarding the Radiated Immunity and Conducted Immunity modules of RadiMation.
Post Reply
incognito
Posts: 236
Joined: 03 Feb 2012, 13:07

Possible to use 4 probes during field calibration?

Post by incognito »

Is it possible to setup a routine for the electric field calibration method with 4 probes for field uniformity test?
User avatar
joro
Posts: 444
Joined: 24 Aug 2011, 09:55
Contact:

Re: Possible to use 4 probes during field calibration?

Post by joro »

Yes, it is possible to use 4 (even up to 8) field probes during the calibration.
There are two different variants to perform a calibration with multiple probes:
  1. Use the average field strength of the 4 probes, as the field strenght that will be calibrated. See the 'One-point calibration' in chapter 12 of the manual. Especially the 'Field Probe Setup' dialog, which allows to select up to 8 probes to be combined in an average value.
  2. Use multiple probes in combination with the constant power method. This will allow to do 4 points of the 16-point Uniform Field Area calibration as once. (See: https://forum.radimation.com/viewtopic.php?p=228#p228 for a detailed explanation)