This version 2024.2.7 has been validated by our internal software quality checks. We strongly advise to test and verify (according to your quality system) this and each new version of RadiMation at for example a separate PC or laptop with your setup and with your measurement equipment, before you are going to use it for daily operation.
Download
The new 2024.2.7 release of RadiMation can now be downloaded from the Raditeq website.

We always advise to thoroughly test each new RadiMation software version to determine if it works correctly in your situation with your equipment, before you use it for daily measurements. This is even a requirement for all ISO 17025 accredited test-labs. To enable customers to validate new software versions before they implement that new software version, we even made it possible to install multiple versions of RadiMation side-by-side on a single PC. It is of course also possible to perform this validation on for example a separate PC or laptop.
Fixed issues
The following issues are fixed in this 2024.2.7 version (compared to the 2024.2.x version):
- #12918: When pressing retry perform a full device driver check
- #14893: Driver: Exodus AMP2070A
- #14894: Driver: Exodus AMP2071A
- #15414: Driver: Rohde & Schwarz UPV
- #15626: Re-activate write termination to use implicit GPIB EOI
- #15670: R&S FPC1500 driver improvements
- #15897: Add hidden member to allow selection of DC high impedance coupling
- #15994: Fix Attachment is not included in report with ||EUT|ATTACHMENT 1 CONTENTS||
- #16036: Send the MPOV command to control the 0-10V DC output of the NX series VDI generator
- #16044: Support the Keysight N9000 series Accelerated Time Domain (ATD) scan mode
- #16073: Improve calibration file frequency range checks and show file name in related errors
- #16107: Support the "Coupling NX5 bs-3-480-32" external surge coupling device
- #16127: NX-series add coupling device bs-3-480-32
- #16139: Narda EMD-100 driver
- #16161: Maturo FCU3.0: add default communication configuration
- #16195: Log memory usage
- #16203: Perform the full device driver checks when the test initialisation is retried on error
- #16216: Questions about which LISN insertion loss corrections should be applied
- #16223: Add specialized Milmega 80RF1000 series amplifier driver without band selectio
- #16223: Improvements for the Milmega amplifier state detection
- #16242: Implemented the missing function GetPointsInTrace for RTE/RTO oscilloscope driver
- #16245: Use correct measure trigger configuration for Keysight 344xx
- #16273: DMM7510 now uses fixed aperture at detector bw 3 or 30
- #16278: Limit the minimal allowed carrier level based on the stepped attenuator option
- #16285: Do not set the changeflag of the limit line file when only opened
- #16291: Allow disabling of resetting the Agilent 33522B when initialized
- #16299: RTE/RTO use STAT:OPER:COND? to detect if the measurement is finished
- #16310: Driver: ETS-Lindgren EMSense 7007-301
- #16312: provide details about the amplifier linearity check
- #16313: Fix Maturo MC32 rotates during measurements without being told.
- #16320: GPIB-Board is not installed or configured incorrectly (GPIB: ENEB)
- #16322: Could not open connection to networked device
- #16325: AR2500A225 amplifier cooling fault raised an interlock error when starting test,
- #16327: IO communication error during test
- #16340: Add pulse modulation measurement support on Agilent N9010A
- #16344: Improve surge sequence progres calculation of the NX series
- #16345: Do not show the identification panel for configurable device drivers
- #16246: Corrected and improved the calculation of power levels during the UFA calculation using constant field 61000-4-20 method.
- #16352: Add delays to the Milmega 80RF1000 series amplifier driver
- #16369: Driver: R&S NRP67S
- #16375: Implement speed control for the Frankonia table and towers
- #16410: Set detector measure time when measuring zero-span receiver mode with analyzer
- #16424: Fix for firmware update RPR4000 series
- #16428: Fixes related to the AR FP7018 manual tuned field sensor driver
- #16445: Do not stage triggered measurement while the zero operation is running when measuring using a VI instrument
- #16488: Fix allow pulse modulated power measurements with the Agilent N9010A
The complete list of release notes can be found on: http://wiki.dare.nl/wiki/index.php/Release_Notes_2024.2.