Hello,
by testing BCI Closed loop, the calculation of the forward power in case the required current cannot be achieved seems to work unpropperly. By testing 100mA with a calibration of 200mA, and a k=4 following seems to happen:
1) Current is not reached;
2) K=4 is applied to the power used for 200mA calibration. In this case i expect, that first the software compares the calibration file with the required current. If the calibration current is not equal to the required current, then a calculation of the needed calibration power assuming a linear system should be done.
3) On the power calculated in Step 2 K=4 should be applied.
Can you check if this is possible? Feel free to cotact me for more information or if you have some questions...
Greetz Marius
BCI Closed Loop
Re: BCI Closed Loop
Thank you for reporting this issue. We have created a new item for this problem with the reference #5218: BCI Closed loop is generating overtesting
Could you please provide us with the following information:
- Send us a copy of the used TSF file
- Which version of RadiMation is used
- Which exact device drivers are used during the calibration (e.g. which jig device driver is used etc.)
- Which exact device drivers are used during the test (e.g. which jig device driver is used etc.)
Could you please provide us with the following information:
- Send us a copy of the used TSF file
- Which version of RadiMation is used
- Which exact device drivers are used during the calibration (e.g. which jig device driver is used etc.)
- Which exact device drivers are used during the test (e.g. which jig device driver is used etc.)